Cyton picking up noise only on one channel?
by JLG ·Using the exact same setup but with the board that has never been used in any of the impedance measurement test: -
Cyton picking up noise only on one channel?
by JLG ·I have in my possession two Cyton boards (let's call them Cyton A and Cyton B ). As reported in another post here https://openbci.com/forum/index.php?p=/discussion/3726/electrode-impedance-test-using- -
electrode impedance test using the GUI widget
by wjcroft ·Richard @retiutut and @philip_pitts, I'm not trying to argue with any one or examine source code differences. The fact is, these two users have proven that the impedance algorithm is busted. Something -
electrode impedance test using the GUI widget
by retiutut ·BoardCyton class for sending the command to start checking impedance: -
electrode impedance test using the GUI widget
by wjcroft ·Richard @retiutut and @philip_pitts, both these users have already found that the resistors they connect to Cyton, do NOT read correctly in the impedance widget. So something must be causing this issu -
electrode impedance test using the GUI widget
by retiutut ·The real source of truth for talking about ADS1299 and injected current impedance measurement is the official Data Sheet. -
electrode impedance test using the GUI widget
by JLG ·When looking at Chip's blog (https://openbci.com/community/openbci-measuring-electrode-impedance/) it seems like he advises using the BIAS electrode. However when looking every other documentation onl -
electrode impedance test using the GUI widget
by wjcroft ·https://openbci-stream.readthedocs.io/en/latest/notebooks/A2-electrodes_impedance.html -
electrode impedance test using the GUI widget
by JLG ·The original version of the impedance check in the old GUI (prior to the widget) used to work properly. So this is likely some issue that surfaced when the code was refactored. I'm guessing that a Git -
electrode impedance test using the GUI widget
by Vedashree ·One other test you could make is with an older version of the GUI, BEFORE the Impedance Widget was introduced. In that era, the impedance measuring was turned on with the little 'ohm' button on each l -
electrode impedance test using the GUI widget
by wjcroft ·JLG, thanks for your comments and confirmation. The original version of the impedance check in the old GUI (prior to the widget) used to work properly. So this is likely some issue that surfaced when -
electrode impedance test using the GUI widget
by JLG ·high peaks at many frequencies and sometimes the impedance values just decrease over time and the ADS ends up railed. -
electrode impedance test using the GUI widget
by wjcroft ·One other test you could make is with an older version of the GUI, BEFORE the Impedance Widget was introduced. In that era, the impedance measuring was turned on with the little 'ohm' button on each l -
Skin contact impedance
by wjcroft ·https://openbci.com/forum/index.php?p=/discussion/3726/electrode-impedance-test-using-the-gui-widget#latest -
electrode impedance test using the GUI widget
by wjcroft ·https://openbci-stream.readthedocs.io/en/latest/notebooks/A2-electrodes_impedance.html -
electrode impedance test using the GUI widget
by Vedashree ·I am measuring Electrode impedance by the method as below, here I am not using RLD (since not testign with head). -
Skin contact impedance
by Vedashree ·Is there a way to measure Skin contact impedance or Electrode tissue contact Impedance? -
Low amplitude in FFT Plot
by Yoshi ·* Impedance: less than 10k -
HIGH IMPEDANCE
by wjcroft ·If you are seeing the same impedance on every channel, suspect your Reference electrode connection. Because with EEG, this is shared as the paired differential pin with each EEG channel. So inspect al -
HIGH IMPEDANCE
by anasofiang ·I´m struggling with my ganglion board because when I make the impedance test, It shows impedance levels of 10737.7k on every channel I´m using. Because of that, I can´t take any measurement, so I wond