Input Current Bias and CMRR degradation
in Ganglion
From the Ganglion design files it appears that the positive and negative inputs of the instrumentation amplifiers are tied to I_BIAS (the midpoint voltage of the INAs) via a 330k resistor.
I'm trying to understand why this was done, because in my simulations the CMRR is much better without them.
If the electrodes were capacitively coupled this would make sense as this would act as the source for the input bias current of the INA. However in my understanding most electrodes are not capacitively coupled, or am I wrong in this assumption?
My understanding is as follows:
First consider a circuit without a DRL or Ground electrode. Just the two electrodes to the INA. This will be unstable because input bias currents will lead to a charge build up or depletion on the capacitors C1 and C2, cause the inputs to go out of the common mode range for the INA.
https://ibb.co/eaOqP9
Adding a “ground electrode” which is driven via an opamp solves this problem. The inputs now stay in the common mode range for the INA as the bias currents have a resistive path to ground. Bonus is that it reduces the effective impedance from the person to ground, lowering the common mode voltage as well.
https://ibb.co/cgBzHUNow looking at the ganglion design, it also adds what I will call the bias circuit. 330k resistors tied to ground just before the input of the INA. If there were no ground electrode this would act as the input bias return path, but we already have that. Secondly it adds resistance in parallel with the INA input, reducing the input impedance and leading to worse quality.
https://ibb.co/dqrMWp
Maybe I’m oversimplifying, but I’d really like to know where my understanding is wrong, or if it really is redundant.
Here are the sims that I ran. (*Note I used a different INA than the ganglion and I omitted the integrator circuit to simplify)
The electrode impedances are 10k and the other 20k, I’ve got the ground electrode at 10k too. The CM input signal is 400mVpp @ 60Hz.
https://ibb.co/dOvXjK
Using the following setup (removing R4 & R5) I get only 0.002mVpp of the 60Hz noise! Quite a difference!
https://ibb.co/fTvXjK