question on uVrms value shown in the GUI
Hello to everyone. I'm working on a project to study the sensibility of the sensors and i've got stuck in a problem. Once i give a known input (Sine: 1 Hz, 4milliVpp, 0offset) i find in the GUI a sinusoidal signal that respect the frequency value and shows a constant value of 1001microVrms.
The problem is that :
Vmax=sqrt(2)*Vrms=1415 microV
Vpp=2xVmax= 2831 microV = 2.8milliV which is not equal to the input of 4 milliV.
Does anyone know why?
P.S. i am using Cyton.
The problem is that :
Vmax=sqrt(2)*Vrms=1415 microV
Vpp=2xVmax= 2831 microV = 2.8milliV which is not equal to the input of 4 milliV.
Does anyone know why?
P.S. i am using Cyton.
Comments
https://drive.google.com/file/d/1KPYdv0c7mD1uZWRPeqSiBwze4vTsJHM5/view
zoom
Rin = 6.8 MΩ
In the meanwhile i'd like to pose you another question, so that when you will answer, i'll be sure and ready on what to do. Do i have to connect the sensors this way: the + signal coming out from the voltage divider connected to the channel pin, and the - signal coming out from the generator (used as reference even by R2) connected to both SRB2 and BIAS?