UltraCortex MK4 - Unreliable readings
Hi,
Thanks
Patrick
I was using a combination of electro-cap and Cyton 32-bit board and everything worked fine; I have also published some papers using this combo.
I decided to upgrade to a UC MK4 headset, however the readings I am getting are unreliable to say the least. Let me explain my setup:
Currently using only 8 channels (even though I have the daisy module for 16 channels if needed). So to keep it simple let's eliminate the daisy module and keep the basic setup of 8 channel.
I have C3, CZ, C4, P3, Pz, P4, O1, O2 as channels from 1-8. I have the BIAS on my left ear (or FPz, tried both) and the ground on my right ear.
I switch on the GUI, and wear the cap and start to tighten the electrodes, until I feel them gently pressing, however most readings where either RAILED, NEAR RAILED or with a high UV (14+) or fluctuating(9-23uV, which is definitely unreliable). With the electro cap I usually had readings ranging from 7uV-14uV, depending on the electrode position and activity at that time, but when idle they where consistent.
Another thing is that when I try to measure electrode impedance from the GUI, on some channels (not on all channels and not always on the same channel, expect channel 1 which is always consistent) it shows a consistent spike in the 31-33Hz range, sort of the 50/60Hz hum you have for electrical interference. Is this normal? When I switch the check impedance off on that particular channel, the 31-33Hz spike goes off.
Honestly I am at a standstill. I have tried tightening the electrodes, but it makes the signal worst. Sometimes I may get the O1 OR O2 to a decent signal, by leaving it running for a few minutes without doing anything.
Any ideas?
Thanks
Patrick
Comments
Let me show you with some screenshots:
Patrick
Thanks for your time.
Patrick
Patrick
Patrick
Sorry for the issue, and sorry for the delay in my response. Can you take a look at your Ultracortex electrodes. If you see any discoloration at the tips of the spikes, it is likely an issue with the electrodes themselves. We do our best to screen for this manufacturing defect but it does happen from time to time. The discoloration will appear as a white-ish color. If you see this, it means that the Ag-AgCl coating has worn off or was not applied thoroughly enough to the ABS substrate.
-Conor
I am also curious about the results you're seeing. I know you tried changing various things, but can you confirm that you switched electrodes/wires from a working channel into a non-working channel? Since I don't see that you stated that specifically, that result might help to narrow things down a bit.
Also, I since designed the geometry of the headset itself, so I am mostly familiar with that aspect, and would like as much information as possible about the context. I have tried the headset on many people in public demonstrations and the biggest variable I'm noticing in results using our dry electrode system related to the subjects hair. There are different electrode comb designs, and picking one that works for everyone is challenging. I hope multiple designs can be offered in the future for optimum performance. If you don't mind, having at least a description of your hair (length/texture/density) might be helpful to me. If you're enthusiastic about that idea picture would be even more appreciated (from the back with headset on would be ok).
Thanks,
-Aaron