EOG cross talk on openBCI channels

swapswap Bekeley
edited January 2017 in Hardware
Has anyone experienced cross talk on the openBCI  (8ch) channels? I think it's coming from the BIAS pin?  pronounced EOG
appears in non-EOG channels only when EOG channels are hooked
up otherwise the impact of EOG is minimal.      Any assistance would greatly be appreciated.

 

Comments

  • wjcroftwjcroft Mount Shasta, CA
    Swap, hi.

    Where is your Ground / Bias electrode positioned? Try moving it to other locations. Similarly where is your Reference?

    Another thing you can do is look at the steps in the tutorial for how to simultaneously measure EEG, ECG and EMG. Since EMG and ECG are such a high level (millivolt vs microvolt) signals -- they will tend to swamp out the weak EEG signals. EOG is essentially the same as EMG. That is why the CHAN SET tab in the GUI is used to disconnect those channels from the EEG.


    William

  • wjcroftwjcroft Mount Shasta, CA
    Note that the scrolling on that link does not always go to the correct location. Use the left hand column to select the V.4. section link.

  • swapswap Bekeley
    edited January 2017
    Hi wjcroft,

    We are doing exactly what the tutorial showing us.

    Let's say we are only concerned with two channels. We have channel 1 N electrode on the outer side of the left eye and channel 1 P electrode connected to the outer side of the right eye.This channel does not contribute to the bias signal generation and SRB2 is disabled.

    Now channel 2 electrodes (N and P) are in front and behind the ear. Again, no contribution to bias signal generation and SRB2 pin off. Just like in the tutorial for the EMG/EKG signal.

    We now go one step further and have no electrodes connected to the Bias pin. Thus, we definately do no have inverted common mode rejection interfering in our measurements.

    What we observe is that the EOG signal picked up in channel 1 "leaks" into channel 2. In channel 2 we see a low pass filtered version of channel 1 at a lower magnitude. If we disconnect channel 1 (physically remove electrodes and unplug pins), channel 2's signal changes considerably. Thus, we believe that the different channel signals are somehow coupled together.

    ... thanks
  • wjcroftwjcroft Mount Shasta, CA
    You must have Bias connected somewhere to give the ADS1299 differential amps a zero reference point.

    If the tutorial works for simultaneous EEG, EMG and ECG, then it should (in theory) be possible for you to replicate that in your situation.

    It could be that your signal leads themselves are coupling by virtue of the close proximity. Conductors that travel along the same pathway induce currents in each other.

    OpenBCI is primarily intended for EEG, and these aux channels require care since they are of such high voltage compared to EEG.

    Just curious, what does in front and behind the ear give you? Is that used in EOG as well?