skin-impedance measurement of active vs. passive electrodes
Hi there,
I'm using the openBCI V3 32bit board.
I've been doing some comparisons between different electrode technologies (wet vs dry, passive vs active).
I have a question regarding the lead-off measurement from the ADS1299.
When using the dry electrode, I'm getting skin impedance of roughly 40 kOhm.
I have active caps which provides high pass filtering and a unity gain buffer.
When I use the same dry electrode with the active cap attached, the skin impedance measurement given by the openBCI GUI is roughly 4 kOhm.
I viewed the trace in matlab after recording it.
Below is the image of the 2 channels when in lead-off detection mode. he top trace is of the passive channel and the bottom trace is of the active channel.
As you can see, the amplitude of the 31.2 Hz signal is much larger in the passive trace, resulting in higher skin impedance.
I assume this is due to the high input impedance of the unity gain buffer, but I don't really know why.
Can someone elaborate on why I'm getting such lower skin impedances with the active cap on the dry electrode?
Thanks..
This paper from EGI,
http://wwe.eeginfo.com/research/ElectrodeImpedance.pdf
Makes clear that 40 K is still excellent for almost all applications. As long as the passive amp has high input impedance, which the ADS1299 does. 1 GigaOhm.
Yes, the paper by Ferree is very interesting and I know that even under 100 kOhm is still quite usable due to the 1 GOhm input impedance of the ADS1299.
I'm more just interested in trying to see the benefits of active over passive in reality. Both Signals look almost identical. Only difference I have found so far is the skin impedance measurement. Even the unfiltered 50 Hz noise is roughly the same amplitude in passive vs active.
Should I be observing improved signal quality with active electrodes (without amplification of the signal, just high pass filter and buffering)?
What characteristics should theoretically be better? SNR? 50Hz interference? Motion artefacts? DC offset? Because at the moment I don't see any significant benefit with the active signal I'm receiving.
Thanks
Nice to see that OpenBCI compares decently with Cognionics dry headset. Do you have an idea of the cost of the Cognionics gear?
This recent paper by Jeremy Frey compares a g.tec amp with OpenBCI. Both amps connected to the same active cap. So it's not an active / passive comparison like yours. But does offer some insights on amp comparison techniques.
https://hal.inria.fr/hal-01328427/document