skin-impedance measurement of active vs. passive electrodes

edited January 2017 in Electrodes
Hi there,

I'm using the openBCI V3 32bit board.
I've been doing some comparisons between different electrode technologies (wet vs dry, passive vs active).

I have a question regarding the lead-off measurement from the ADS1299.
When using the dry electrode, I'm getting skin impedance of roughly 40 kOhm.
I have active caps which provides high pass filtering and a unity gain buffer.
When I use the same dry electrode with the active cap attached, the skin impedance measurement given by the openBCI GUI is roughly 4 kOhm.

I viewed the trace in matlab after recording it.
Below is the image of the 2 channels when in lead-off detection mode. he top trace is of the passive channel and the bottom trace is of the active channel.
As you can see, the amplitude of the 31.2 Hz signal is much larger in the passive trace, resulting in higher skin impedance.

I assume this is due to the high input impedance of the unity gain buffer, but I don't really know why.
Can someone elaborate on why I'm getting such lower skin impedances with the active cap on the dry electrode?

Thanks..


image

Comments

  • wjcroftwjcroft Mount Shasta, CA
    edited June 2019
    I would suspect the buffering in the amplifier co-located at the electrode. I'm not sure the Texas Instruments technique for passive electrode impedance measurement is applicable to actives, which are considered superior to passives. But at a huge cost increment.

    My other hunch is that the value '4K' is suspiciously 1K down from the magic number '5K', which has been touted as the "research standard" for impedance measurements. (From a couple decades in the past, with older amplifiers.) In other words, the buffer amplifier electronics output might be juggled to produce a value that people 'expect' to see for a stellar skin impedance value. It might not reflect reality.

    This paper from EGI,

    http://wwe.eeginfo.com/research/ElectrodeImpedance.pdf

    Makes clear that 40 K is still excellent for almost all applications. As long as the passive amp has high input impedance, which the ADS1299 does. 1 GigaOhm.

    William

  • edited January 2017
    Thanks for the reply William.
    Yes, the paper by Ferree is very interesting and I know that even under 100 kOhm is still quite usable due to the 1 GOhm input impedance of the ADS1299.

    I'm more just interested in trying to see the benefits of active over passive in reality. Both Signals look almost identical. Only difference I have found so far is the skin impedance measurement. Even the unfiltered 50 Hz noise is roughly the same amplitude in passive vs active.

    Should I be observing improved signal quality with active electrodes (without amplification of the signal, just high pass filter and buffering)?

    What characteristics should theoretically be better? SNR? 50Hz interference? Motion artefacts? DC offset? Because at the moment I don't see any significant benefit with the active signal I'm receiving.

    Thanks
  • edited January 2017
    Also, on a side note..
    In Ferree's paper, the equation on page 5 for potential difference of the signal received by the amplifier..
    I seem to come out with much lower signal losses than what Ferree states.
    Ignoring the amp impedance mismatch as he does and ignoring the last (3rd) term as I assume (not sure) he does.. You're left with:

    (Va - Vb) = ((V1 - V2)/2) * (2 - (Z1 + Z2)/Zin)
    Taking Z1 = Z2 = 50 kOhm and Zin - 200 MOhm, I get the following result..
    (Va - Vb) = ((V1 - V2)/2) * (2 - (50000 + 50000)/200000000) = 0.99975.

    So a max signal loss of 0.00025%... not 0.025% as in he paper.

    Any idea?
  • wjcroftwjcroft Mount Shasta, CA
    Mark, hi again.

    re: benefits of actives over passive. It appears generally that these full cap active systems and amplifiers cost thousands of dollars and are only affordable to clinical and research groups. In that arena, especially in noisy operating rooms or labs, the increased noise immunity of the actives may be the most prominent benefit.

    It sounds like you are just outputting microvolts from your actives, the same signal levels as passives. Is it possible that some active systems output higher signal levels that are internally converted back down inside the amp? This higher signal level on the cables would likely provide better immunity to noise and movement. Some actives also use shielded cables.

    Since I have no experience myself with active systems, I would defer to your manufacturer. Can you say what you are using?

    Actually OpenBCI itself might perform better in terms of noise factors with something like the Electro-cap / gel 10-20 caps. The wet skin interface always is more stable because the electrolyte layer is less pressure sensitive.

    re: math in Ferree's paper. I believe he is still at EGI or UOregon, you could write him there.

    Regards,

    William
  • Yes, same signal levels as passive. The method of transmitting at high voltage over the wire and reducing back to original voltage at the front-end isn't used.
    The cap simply implements a highpass filter and unity gain buffer.
    The active cap is from Cognionics.

    Below is a image of a few plots from Matlab of the dry active channel vs dry passive channel.
    Taken from Fp1 with reference at A1 and ground at A2.
    image
    The sections at 11-15 secs, 42-46 secs, 70-74 secs, 92-96 secs are blocks of rapid eye blinks just to give an idea of the amplitude you're looking at.
    There are some slight differences, but probably not enough to warrant the extra costs involved with active.

    P.s. thanks, I actually found a paper by Ferree on his UOregon site discussing the maths. :-)

  • This is a better view of the FFT data.
    image
  • wjcroftwjcroft Mount Shasta, CA
    Mark, thanks for those graphs!

    Nice to see that OpenBCI compares decently with Cognionics dry headset. Do you have an idea of the cost of the Cognionics gear?

    This recent paper by Jeremy Frey compares a g.tec amp with OpenBCI. Both amps connected to the same active cap. So it's not an active / passive comparison like yours. But does offer some insights on amp comparison techniques.

    https://hal.inria.fr/hal-01328427/document

    William

  • That's not exactly comparing Cog vs OpenBCI. Just comparing two openBCI channels, one with a standard dry sensor, one with the dry sensor connected to an active cap.
    Thanks for that paper. Nice to see openBCI competing well!

    Just to clear up the skin-impedance measurement in my head.
    Is the 31.2 Hz, a 6nA signal? Passed through the REF to the skin and then picked up and measured on the "recording" channels? Or am I getting mixed up?
    Looking at the openBCI GUI, when impedance checking, I see something like this:

    image


    (V*sqrt(2)) / I = ( 76.4u * 1.41 ) / 6n =  18 kOhm.

    Is this value, the skin-electrode impedance PLUS the 5 kOhm series resistor? I would ahve thought the OpenBCI GUI would display it without the 5 kOhm?

    Mark 