sequential (bipolar) montage wiring setup
Hi,
I got my Open BCI 32bit board and headset. I managed to connect everything up and successfully setup the connection as 'Referential Montage' and I get the signals I need. However I am at a loss on how to setup the connections on the board for a Bipolar Montage. Anyone got any tips? I am looking for:
This question might be simple, but it's mixing me up.
I got my Open BCI 32bit board and headset. I managed to connect everything up and successfully setup the connection as 'Referential Montage' and I get the signals I need. However I am at a loss on how to setup the connections on the board for a Bipolar Montage. Anyone got any tips? I am looking for:
1. Cz as reference (C3-CZ, C4-CZ etc)
2. Longitudinal Bioplar (Fp1-F3, F3-C3, C3-P3, P3-O1, Fp1-F7 etc).
Thanks
Comments
1. if you just want everything referenced to Cz, put Cz on SRB2 and use the same default setup as the tutorial.
2. You will need to make some Y-cables. And use the CHAN SET tab to disconnect the channels from SRB, as in the ECG, EMG part of the tutorial.
Your Y cables can be made from short jumper cables such as,
https://www.adafruit.com/products/1951
https://www.adafruit.com/products/1956
You will be cutting these in half, striping the ends about 1/2 inch, and making the Y's. Each Y consists of two females and one male. Twist all 3 wires together (possibly soldering for best connection), then use shrink wrap or tape to insulate the bare wire ends.
An example wiring setup might look like:
Fp1 - IN1N
F3 - IN1P using Y to bridge IN1P and IN2N and male going to F3 electrode
C3 - IN2P (similar idea with Y to IN3N)
P3 - IN3P (with Y to IN4N)
etc.
William
EEG inputs are always differential with + and - pairs. With the default referential setup SRB2 is the common + bus. The IN*N pins are the electrodes.
Some medical fields demand the sequential montage, such as,
https://www.google.com/search?q=sequential+montage+epilepsy
http://erpinfo.org/erplab/erplab-documentation/tutorial_4/Channel_Operations.html
I would not know if this kind of derived montage would be acceptable in clinical settings that require greatest accuracy, for example.