Can anyone explain about how ganglion measure the electrode impedance? And is ganglion measure impedance for a different frequency or for a certain frequency?
Thank you and have a happy Christmas and new year...
If you examine the code section linked above, you can see that the injected waveform is NOT a pure sine wave, which is a single frequency. But instead a triangle wave is injected, which carries a range of frequencies. But actually EEG waveforms are not sine waves either, and frequently have 'spiky' / jagged waveforms.
I have another question. For my work, I need impedance values during the measurement. If I record the impedance value from the software interface will it be accurate or do I need further calculation to get the impedance across the electrode? Some of the publications they use this formula: Actual Average Impedance(Ω) = Average Impedance(Ω) − 5000 to calculate cyton electrode impedance.
By injecting current into a channel to measure impedance, it interferes with the actual signals that can be picked up from the electrode. Also, this can only be done on one electrode at a time, since there is signal bleed from injecting the current on a given channel into other nearby channels. For Cyton, you could check impedances first, then use a "Railed Percentage" calculation to roughly gauge the signal quality from all electrodes at once.
Thank you for the reply. I am recording the skin impedance from GUI for the Ganglion device. Is the GUI impedance data is raw data of that particular channel or do I need further calculation to get the skin impedance? As I got an answer from the code that it uses a 4 kHz single frequency during impedance measurement. And is this impedance value is reference value-dependent of the Ganglion?
The GUI widget is showing skin impedance. Check the widget source to see if it is applying more calculations, beyond what the firmware is doing.
I don't believe this code section (previous link repeated below) is injecting 4 kHz. It is a much lower frequency than that. Frequency is in the brainwave region, no higher than 40 Hz. As I mentioned before it is a triangle wave, so not a single frequency.
Musfequr, hi.
See the multi-page schematic here,
https://raw.githubusercontent.com/OpenBCI/Ganglion_Hardware_Design_Files/master/Ganglion_SCH.pdf
Impedance injection DAC is located on the 2nd page. And then see the impedance code section here:
https://github.com/OpenBCI/OpenBCI_Ganglion_Library/blob/master/OpenBCI_Ganglion_Library.cpp#L404
A fixed frequency AC signal is injected, then resistance measured via ohm's law: R = voltage / current.
https://en.wikipedia.org/wiki/Triangle_wave
Thank you @wjcroft for your reply. I will be helpful for my work.
https://github.com/OpenBCI/OpenBCI_GUI/blob/master/OpenBCI_GUI/W_GanglionImpedance.pde
Thank you william