Noise and huge offset with Cyton and new gold cup electrodes [resolved]

iswapneeliswapneel Bangalore, India
edited November 2020 in General Discussion

I have recently got Cyton Board from a friend and while trying a tutorial from

https://docs.openbci.com/docs/01GettingStarted/02-Biosensing-Setups/EEGSetup

Signal I got on GUI was noisy and with huge offset. I tried couple of experiments
1. Short Channel 1 + SRB2 + Bias : EEG signal Range 4000uV+
2. Short Channel 1 + SRB2 with bias left dangled : EEG signal range ~10uV
I have uploaded photos of GUI and board setup here

https://drive.google.com/drive/folders/1lxzXB4JnN5DRvpIsIuO8Vo6q8sTu7Vif?usp=sharing

I was expecting signal to be in range of 20uV but I am getting signal in range of 10000uV+ in experiment 1. I tried with another channel and got around 4000uV+ signal. I am not sure what is the issue with the setup.

Would really appreciate any help.

Comments

  • wjcroftwjcroft Mount Shasta, CA

    Swapneel, hi.

    When was this board purchased? Did it EVER work correctly, for example, with your friend?

    Shorting all three electrodes: channel, SRB2, and Bias, SHOULD result in close to zero microvolts. If that is not happening, then something is wrong. Have you tried swapping some of your electrodes around? A gold cup electrode, if chipped, will expose the tin metal underneath, creating a battery effect. Are these gold cups from OpenBCI? What paste are you using?

    Regards, William

  • iswapneeliswapneel Bangalore, India

    Hi William,
    Thanks for your reply ! Electrode are from OpenBCI and they have arrived a week ago.

    I tried few more experiments and found that culprit was the bad conductive paste. With exact same setup and with new conductive paste (Ten 20 ) I am getting expected EEG signal.

    Cheers,
    Swapneel