is continuous monitoring of impedance possible while recording EEG?
Hello!
Is continuous monitoring of impedance possible while recording EEG?
I guess this is not possible with the OpenBCI, however, are there systems which do this?
Or do researchers lower impedance as much as possible at the start of a recording session and perhaps check again at the end of a session to see how it changed?
Thanks! k
    The input impedance of the ADS1299 drops from 1 gigaohm to 500 megohms with the ADS1299 during impedance measurement. And also the current impedance algorithm injects a 30.x Hz signal (nanoamp range), so cannot be simultaneous with data acquisition.
    Some EEG amps do provide continuous monitoring. These tend to be higher end systems.
    I replied on the other thread regarding acceptable impedance ranges.
    Great, many thanks for your reply William.
    k