Self made electrode shows "RAILED"
in Cyton
Hello,
I have been experimenting with making my own set of electrodes for EEG measurement. I prepared two set of electrodes with different set of materials. The first type of electrode is working fine and giving the desired EEG signals. However, when I connect the second type of electrode (SRB pin still connected to the first type of electrode) , the signals are "RAILED". If I hold it firmly with my hands, it does go out of the RAILED state but does not give good EEG signals.
I have checked that the impedance of both the electrode is same, at around 1 ohms.
What other electrical properties or requirements am I missing?
I have also tried the following things to debug my electrodes -
1. Connected SRB electrode and N1P pin electrode and got a 0 uV output.
2. Connected the electrode with a sinusoidal signal provided through AFG (with SRB connected to the ground of the AFG). I got the desired peak in the FFT in the OpenBCI GUI.
What can I do in order to ensure a continuous and good quality signal?
I have been experimenting with making my own set of electrodes for EEG measurement. I prepared two set of electrodes with different set of materials. The first type of electrode is working fine and giving the desired EEG signals. However, when I connect the second type of electrode (SRB pin still connected to the first type of electrode) , the signals are "RAILED". If I hold it firmly with my hands, it does go out of the RAILED state but does not give good EEG signals.
I have checked that the impedance of both the electrode is same, at around 1 ohms.
What other electrical properties or requirements am I missing?
I have also tried the following things to debug my electrodes -
1. Connected SRB electrode and N1P pin electrode and got a 0 uV output.
2. Connected the electrode with a sinusoidal signal provided through AFG (with SRB connected to the ground of the AFG). I got the desired peak in the FFT in the OpenBCI GUI.
What can I do in order to ensure a continuous and good quality signal?
Comments
Thanks for the answer.
I now understand the reason for the signals getting railed. Is there any suggestion on how can I solve this situation?
Should I use same kind of electrode (both SRB pin and N1P pin)? Will that help in getting non Railed Signal?
Any other advices in this regard will be appreciated. Thanks!!
@wjcroft
1. It is basically a highly conductive cloth of Copper (Cu) and Nickel (Ni)
2. The other electrode connected to SRB is silver chloride electrode, and it works when two of such are used to take measurements. I now understand that this may be causing me additional troubles.
I am trying to procure Silver coated cloth electrode for my application.
Meanwhile I tried applying silver conductive paste on top of the above mentioned Cu and Ni cloth electrode. Although the surface is fully coated with the silver conduction gel but still I am facing the same problem.
Can you suggest if silver conductive paste on top of copper conductive cloth should work or not?
If it should, then I am having more issues with my setup.