What is this shift in the baseline?
by wjcroft ·re: Nyquist limits of bandwidth -
Calculating PSD with 16-electrodes OpenBCi
by Soroushzrzz ·Hz) bands. By resampling to 60 Hz, which has a Nyquist frequency of 30 Hz, I can accurately capture these critical frequency ranges. Additionally, resampling reduces the data size and computational l -
EMG bandwidth to calculate Shannon–Hartley theorem
by wjcroft ·EMG in labs sometimes use higher sample rates, for example 500 Hz or 1000 Hz. However with Ganglion you are limited to 200 Hz sample rate. Technically by the Nyquist rule that would limit you to wavef -
FFT Data Transmission via LSL with Ganglion Board
by Otemo ·However, this configuration seems to exceed the Nyquist frequency of 100Hz for a 200Hz sampling rate, as 0.9765625Hz x 125 equals approximately 122Hz. Could you please provide insight into how the FF -
F0 coding of speech-like stimuli using the OpenBCI EEG recording
by wjcroft ·If your signal is in the range of 100 to 126 Hz, then a 250 Hz sample rate is not going to do well at all. 'Technically' the Nyquist limit says a 250 Hz sample rate can record a 125 Hz signal, but tha -
on open pins (not connected to anything), Is this signal normal?
by jnvandermeer ·is quite good at picking up anything below its Nyquist frequency that is around. Probably if you go to an electronically 'quiet' area, like in the desert or uninhabited mountain area, you will get a -
Noise at Nyquist Frequency (125Hz) [resolved]
by wjcroft ·125 Hz EEG from the scalp, and samples at the Nyquist frequency (125 Hz) do not make any sense because of aliasing effects. You generally want 4 to 5 samples per waveform: 250/5 = 50 Hz max. If only -
Noise at Nyquist Frequency (125Hz) [resolved]
by JTBME ·Hi! I've been experimenting with independent referencing with tests using 8 channels at fs of 250Hz. The first 6 channels are EEG, 7 is empty, and 8 is EMG. When I plotted the raw data in Matlab, I … -
FFT Plot Background
by retiutut ·It depends on the sampling rate and follows https://en.wikipedia.org/wiki/Nyquist–Shannon_sampling_theorem -
Artifact pattern on WiFi Shield + Cyton [resolved]
by wjcroft ·Josh (strikeroot) has suggested that due to the Nyquist theorem, he could achieve 125 Hz EEG with 250 Hz sample rate. That is not true and has been explained on previous forum posts here, search on N -
GUI "No such file or directory" error, on Fedora
by wjcroft ·You did not say which board. Cyton IS "DC coupled", but Ganglion is not. Top end is dependent on the sampling rate. Nyquist recommends two samples per waveform, but that is subject to aliasi -
How to change the sample rate in openBCI GUI?
by wjcroft ·A disadvantage to down sampling at the rates you suggest, is that you will lose most of the higher brain wave frequencies. Due to the Nyquist limit. Practically, you want four or so samples per EEG wa -
bandwidth selection in between 50 Hz to 500Hz
by wjcroft ·The Cyton sampling rate without Wifi Shield is 250 Hz. This gives a bandwidth according to Nyquist limit, of 0 to 125 Hz. But in actually, more like 0 to 50 Hz, since you want multiple samples per wav -
Questions on the Ultracortex Electrode
by wjcroft ·Frequency is determined by the sample rate, according to the Nyquist limit. Typically you want about 4 or 5 samples per waveform. So 250 / 5 = ~ about 50 Hz max. Nyquist would say divide by two, 250 / -
have new Ganglion, want to develop custom firmware for feedback
by logical ·You have to get 2x the maximum frequency for the nyquist sample rate. I've done some pretty good stuff on processors with similar specifications. -
Ganglion BLE dropped packet IDs (with Android library)
by wjcroft ·Can you see that sample rates anywhere CLOSE TO the Nyquist limit are going to give extremely aliased and poor reproductions of the sampled signal? Look at some of the diagrams in that white paper. Su -
Ganglion BLE dropped packet IDs (with Android library)
by fwende ·re Nyquist -
Ganglion BLE dropped packet IDs (with Android library)
by wjcroft ·re: Nyquist and aliasing. Do you understand the example I gave with the 50 hz samples of a 100 hz waveform? It will be hugely and badly distorted if sampled close to the zero crossings. The only way t -
Ganglion BLE dropped packet IDs (with Android library)
by wjcroft ·Nyquist limit gives terrible aliasing. Imagine say a 50 hz sine wave that we are sampling. If you are lucky to sample it at the + and - peaks (very lucky indeed), then your 100 hz sample rate gives yo -
Ganglion BLE dropped packet IDs (with Android library)
by fwende ·However if we are looking at 100sps we should get to roughly 45 Hz with Nyquist + 10%, which is the practical minimum iirc. What is your reasoning for 25Hz?