Ganglion 4-channel connection issues and high electrode impedance :(
by matildee3 ·* Electrode impedance -
Identical FFT and Time-Series Signals Across Channels on Cyton + Ultracortex Mark III
by nugroho_budi ·That reading doesn't really tell you what you want, and 4–20 MOhm between SRB and BIAS is expected — don't worry about it. SRB and BIAS are separate nets and aren't supposed to be connected to each ot -
Identical FFT and Time-Series Signals Across Channels on Cyton + Ultracortex Mark III
by danysab ·That reading doesn't really tell you what you want, and 4–20 MOhm between SRB and BIAS is expected — don't worry about it. SRB and BIAS are separate nets and aren't supposed to be connected to each ot -
Identical FFT and Time-Series Signals Across Channels on Cyton + Ultracortex Mark III
by danysab ·The giveaway is in your screenshots: all active channels sit at ~57 uVrms and go "Near Railed" together, while the impedance check shows finite values (83–466 kOhm) — so the inputs aren't fl -
Unipolar signal recording on plants
by wjcroft ·What you are seeing with the same amplitude on different leaves, is likely because the 'impedance' or surface resistance in ohms of plant tissue, is very low. The plant cells, filled with water and di -
high impedence
by gg_022006 ·yes, we did because the jumper cable have excessive width we got some male to female jumper cables from our robotics lab and connected to that our cyton then the male side to the female side of the op -
high impedence
by wjcroft ·Please read this comment regarding acceptable Cyton impedance ranges, -
high impedence
by gg_022006 ·Thank you for your advice. We tried it, and it reduced our impedance to below 100 kohms. -
high impedence
by gg_022006 ·* In the OpenBCI GUI impedance/N-status check, we typically see around ~5000 kΩ impedance per channel even after all the gel application. -
wanted to buy: Cyton + Daisy 16ch — stroke rehabilitation research
by GregBulannikov ·Happy to request impedance test video before -
All electrodes railed (Daisy device)
by lweberairbus ·I'm using dry electrodes. They definitely often have poor skin contact, but I already tested them on a person without any hair and had the same outcome at one point. Also, the impedance test is workin -
All electrodes railed (Daisy device)
by lweberairbus ·All signals are railed, although the impedance test is confirming low resistivity. -
Lower SNR after first day of use
by smanne ·I'm using the electrode gel cap with the Cyton Board and while I got good SNR on the first day of use, it has dropped significantly on subsequent days and I can't seem to recover it. After the first d -
What caused this abnormal (high frequency noise) phenomenon?
by Zenton ·When analyzing data using EEGLab, I noticed increased amplitude and strong high frequencies in channels F3 and F4. I then checked the impedance values of channels F3 and F4, which were normal, aroun -
White film on gelfree electrodes – cleaning and maintenance advice
by wjcroft ·Are you still seeing expected impedances? In other words has the impedance degraded over time, since they were new? -
White film on gelfree electrodes – cleaning and maintenance advice
by JAraneda ·* At what point should this be considered normal wear vs. electrode degradation affecting impedance and signal quality? -
Thinkpulse: signal quality, default settings & impedance ?
by wjcroft ·By Good Signal vs Noise metrics- I meant, is there any objective measure (like impedance) to confirm the contact quality so we can say “this channel contact/setup is OK and we are getting a good sign -
Thinkpulse: signal quality, default settings & impedance ?
by uditkumar007 ·By Good Signal vs Noise metrics- I meant, is there any objective measure (like impedance) to confirm the contact quality so we can say “this channel contact/setup is OK and we are getting a good sign -
Thinkpulse: signal quality, default settings & impedance ?
by wjcroft ·The GUI / Cyton 'impedance widget' WILL NOT work with Thinkpulse, because that widget depends on the ADS1299 chip acting as current source on the passive (non active) electrode having direct skin cont -
Thinkpulse: signal quality, default settings & impedance ?
by uditkumar007 ·In the discussion above, we saw that Julien mentioned the impedance test should be correct with a x24 gain, but he didn't test it. We tested it with "no contact" with the head and it doesn't