EEG Single-channel minimum electrode montage
I am trying to re-design the Cyton, for a specific application which only requires a single differential channel of EEG.
I want to design a board with as little components and as small a footprint as possible.
My main issue is, I hope to only use 2 electrodes to record a differential channel of EEG (for example C3-C4) and I'm not sure if that is possible.
To test this I used the openBCI Cyton with an arbitrary waveform generator (AWG) outputting +/-50uV EEG signals.
I connected the output of the AWG to 1P. I connected the ground of the AWG to 1N and to Bias (essentially shorting 1N and Bias).
In the openBCI software, I included Chan1 in Bias and turned off SRB.
It worked perfectly and I got a correlation of 0.99 between the openBCI data and the original AWG data.
I modified Fig. 72 from the ADS1299 datasheet to show how I intend to setup the channels.
Is this possible? If so why hasn't everyone been doing it? I have some theoretical reasons why it might not work.
Firstly, if I short 1N with Bias, then the - input of the differential amplifier will no longer have the common-mode signals, as they will be inverted and summed with 1N, and cancelled. But 1P will still have them, so I'm effectively ruining the common-mode rejection?
Secondly, ground is usually on the ear as there is no EEG source signals on the ear lobe, only the interference signals (mains, EMI, etc.), which we want to be cancelled out.
Whereas, for example if the signals from C3 and C4 are summed, inverted and injected back into the patient through C4, then interference signals will be cancelled out, but also potentially the EEG signal sources close to the electrode at C4?
Has anybody experimented with such minimalist electrode montages? Is it possible to short ref and bias?
If not, then is it possible to place 1P at C3 and then separate 1N and bias electrodes very close together at C4? Will I still record the equivalent of C3-C4, without diminishing or corrupting the signal?
Thanks
Mark
a differential capacitor CFIlt that shunts the inputs rather than individual RC filters whose capacitors shunt to
ground. The differential capacitor configuration significantly improves common-mode rejection because this
approach removes dependence on component mismatch."
https://e2e.ti.com/resized-image/__size/700x0/__key/communityserver-components-multipleuploadfilemanager/079a2162_2D00_01b9_2D00_402d_2D00_a907_2D00_741417b5e96c-347099-complete/Schematic-ADS1299-2-electrode-V5.png