Active Electrode Circuit Design

in Electrodes
Hey,
Starting a thread here about the best way to go about designing active electrode circuitry. From what I understand, placing an active (op amp or other) device in very close physical proximity to the contact site improves noise reduction and allows for a dry contact between the electrode sensor (gold, silver, Ag-Cl, etc) and the subject scalp.
My basic reference for this is the active electrode circuit published by Olimex
I've used these, and they perform quite well with OpenBCI (5V differential between AVDD and AVSS to power).
The basic shape of the thing has
TVS and Diode suppression on both ends
Power supply decoupling and zener protection on the V input
Some impedance added to the input and output
High Frequency rejection filter on the op amp
but beyond that, it is just a unity gain follower.
Of course, the specs of the op amp are going to play a very important role:
high impedance inputs
high CMRR
low equivalent input noise at low frequencies
My question is, what about the basic circuit, besides the op amp selection ('tho that could enter this discussion) stands to be improved? Or discarded?
Comments
Section 3.3.1 in this paper I posted on the Dry Electrode thread, they are using a modified OpenEEG circuit. And this is Imec no less, one of the leaders in EEG tech.
http://www.mdpi.com/1424-8220/14/12/23758/htm
The schematic is there as well as the small circular board layout. They put the tiny circuit board on the back of their EPDM rubber combs. Dang, this looks pretty easy!
Here's a high rez image of their circuit board you can zoom into. Open in another tab in your browser, then zoom in.
http://www.mdpi.com/sensors/sensors-14-23758/article_deploy/html/images/sensors-14-23758f4.png
The TI chip on that tiny board is, http://www.ti.com/product/tlc2272
It has very good EMI immunity as well, eliminating extra components,
http://www.ti.com/lit/an/sbot014/sbot014.pdf
---
A few more:
Here's an old paper on BioSemi's actives,
http://users.dcc.uchile.cl/~peortega/ae/
I see @jarek had a hand in this as well. Jarek do you have any comments on the circuit board that Imec is using (image link previous post), or the TI 2272 op amp? This all looks pretty sweet and doable without custom ICs. The Imec design seems more compact than the Olimex.
yes, I spent a long time researching AE after I built my first EEG device (OpenEEG). Looks like Imec's circuit is very similar to my first prototype.
I am not an electrical engineer, so I can't comment here on any electrical properties. But those AE I built worked A LOT better than passive electrodes used with my OpenEEG device (old version I soldered myself, without SMT) which required shielded wires.
Since then however I found that other (commercial) amplifiers offered quite comparable (to what I measured) results using just passive electrodes (even without skip preparation, placed on forehead under a headband).
So I am not using AE in my training. They are bulkier (harder to setup) and very sensitive to movement.
Also I think the AE I built may have broken eventually. So the improvement (proposed by Joerg and now sold by Olimex) may be well justified. As I recall its primary purpose was ESD protection.
I remember there was one more idea quite recently which I found absolutely fascinating - capacitive AE (it was discussed on openeeg email list). But I am not sure that is doable. Some people (like Stefan Jung) spent some time on that with no luck.
Jarek, I absolutely agree that modern amps like the ADS1299 produce excellent data from passive sensors. Whether used with paste, gel, or saline solution pads.
Where all the consumer device direction seems to be going these days are dry + active sensors. The dry part is almost a requirement if you want to have a headset that you don like a cap. And since dry sensors have a more tenuous and varying impedance connection to the skin, they are frequently paired with some kind of active circuitry for best signal quality.
Over on this other thread I posted some photos of the Imec EPDM conductive rubber comb sensors. That their active board mounts on the back of.
http://openbci.com/forum/index.php?p=/discussion/138/buying-dry-active-electrodes#latest
Chip will be receiving soon some Ag-AgCl small passive comb sensors. (Links are on that page.) It will be interesting to see how those perform purely passively. And then to imagine or try out some kind of active augmentation if it seems beneficial.
The Imec paper cited above seems to conclude that the rubber comb must be used with the active, otherwise they don't get the results similar to wet passive sensors. Maybe this is just a characteristic of these comb style sensors, I don't know.
The rubber combs I'm sure will be more comfortable than pointy metal combs pressed against the skin. But the comb approach seems a popular one, being used in a number of commercial designs.
Then again there are the dry active polymer sensors used in the new Emotiv Insight. This is not a comb at all, just a smooth rounded bump. But their headset is designed to be more or less "slid" into the hair as you would a comb. Not pressed on from outside like more typical headsets.
William
http://www.gtec.at/Products/Electrodes-and-Sensors/g.Electrodes-Specs-Features
Notice how all their actives just use TWO leads(!) [Touchproof connector has two contacts.] This supplies both power going to tiny active board -- AND must also be the return for the amplified signal. In other words, they must be doing some kind of modulation on top of the DC to get that information back to amp.
@jfrey , do you have any idea what they are doing here, since you have these amps in your lab. Karl @kzurn is testing an active board of his FRI design, but I think that is using 4 leads.
William
http://www.ncbi.nlm.nih.gov/pubmed/16485759
I'm bad at electronics so I don't know what it's worth...
Found the pdf file as well. Cool.
William
The active sensor tiny board is supplied with a constant voltage over the 2 wire DC supply pair. THEN the sensor board adjusts it's current draw based on an translation / amplification of the microvolts measured.
Meanwhile, back at the power supply, the changes in current draw are translated back into voltage changes, which are then relayed to the ADC / amp.
A clever dance that results in 2 less wires. But significantly more complexity at the amp side.
If you look at the way g.tec did their g.SAHARA active dry combs, they have a separate box (g.SAHARAbox) that does this translation. The output of that box goes into their amp which can then be either passive or active sensors.
William
Matsuzaka, Y., Ichihara, T., Abe, T., & Mushiake, H. (2012). Bio-amplifier with Driven Shield Inputs to Reduce Electrical Noise and its Application to Laboratory Teaching of Electrophysiology. J Undergrad Neurosci Educ., 10(2), 118–124. Retrieved from http://www.ncbi.nlm.nih.gov/pubmed/3598091
Gargiulo, G., Bifulco, P., Calvo, R. A., Cesarelli, M., Jin, C., & van Schaik, A. (2008). A mobile EEG system with dry electrodes. In 2008 IEEE Biomedical Circuits and Systems Conference (pp. 273–276). IEEE. doi:10.1109/BIOCAS.2008.4696927
Gargiulo, G., Bifulco, P., Calvo, R. A., Romano, M., Ruffo, M., & Shephard, R. (2011). Giga-ohm high impedance FET input amplifiers for dry electrode biosensor circuits and systems. In K. Iniewski (Ed.), Integrated Microsystems: Electronics, Photonics, and Biotechnology (pp. 165–194). CRC press. Retrieved from http://www.researchgate.net/publication/255994293_Giga-ohm_high_impedance_FET_input_amplifiers_for_dry_electrode_biosensor_circuits_and_systems
Usakli, A. B. (2010). Improvement of EEG signal acquisition: an electrical aspect for state of the art of front end. Computational Intelligence and Neuroscience, 2010, 630649. doi:10.1155/2010/630649
It is something I *also* planned to try out, but then I spent more time on the software than on the hardware. It seems to require "only" a bit of soldering and to choose the right cable (coax?).
http://www.ncbi.nlm.nih.gov/pmc/articles/PMC3598091/
It looks like the "driven shield" is a feature of the Burr Brown instrumentation amp they chose, I'll do some more looking into that part number. This idea is also mentioned in the Fraden text (below).
integrated circuit (IC) is a differential amplifier: i.e., it takes
inputs from the reference and the indifferent electrodes (pins 6 and 3,
respectively) and amplifies their difference, thus subtracting out the
noise common to both inputs. A notable feature of this IC is the “driven
shield” inputs: i.e., it holds the shield of the input coaxial cable at
the same voltage as the electrodes connected to the input through the
buffered guard drive pins (pins 2 and 4 for negative input pin, pins 5
and 7 for positive input pin). As a result, the capacitance between the
electrode and the shield is cancelled, thus preventing the electrostatic
interference through the capacitive coupling between them (Fraden, 2003).
William
https://drive.google.com/file/d/0B8Z-FI2dBPKKU3hzX3lXc09yNW8/view?usp=sharing
And the text describing the figure. Preceding text mentions active shielding board traces; then that is generalized to driving the sensor coax shield:
Figure 5.4b shows a voltage follower connected to the inverting input of an amplifier. The follower drives the shield of the cable, thus reducing the cable capacitance, the leakage and spurious voltages resulting from cable flexing. A small capacitance at the follower’s noninverting input improves its stability.
More background info,
https://www.google.com/search?q=driven shield
http://en.wikipedia.org/wiki/Driven_guard
http://www.nvk.com.tw/eeg-lead-wires/active-eeg-electrodes.html
William