How to measure impedance using Cyton serial command protocol
edited January 2020 in Cyton
We can measure the impedance of each electrode using OpenBCI_GUI.
I want to measure it from my software, using Cyton serial protocol.
According to the SDK, there is a command to turn on/off the Impedance Setting mode.
https://docs.openbci.com/docs/02Cyton/CytonSDK#leadoff-impedance-commands
If I set the electrode 1 to the Impedance Setting mode, then what should I do after that?
    edited April 27
    Hi Hiroki,
    See this section in the GUI Processing code:
    https://github.com/OpenBCI/OpenBCI_GUI/blob/master/OpenBCI_GUI/DataProcessing.pde#L52
  • Let me confirm the calculation.
    I think I should send 'z 1 0 1 Z' to cyton since I use N pin to measure, is this right?
    After I receive the EEG data, I should multiply the data with -1, since the data is negative upper, is this right?
    When I turn the impedance check mode on using OpenBCI_GUI, the impedance is shown as around 640kOhm.
    But my program which measures impedances shows that it is around 2900.
    My program calculates impedances according to the above code, like: sqrt( 2.0 ) * (-v) * 1.0e-6 / 6.0e-9 - 2200 ( v is the EEG data streamed ).
    Something wrong with my program. Do you have any idea?
  • Hi William,
    You might add [resolved] to this thread title but is not resolved.
    Can you provide additional information?
    edited January 2020
    // Subtract the 2.2kOhm resistor
            impedance -= cyton.get_series_resistor(); // Subtract 2200
    If you say your program outputs around 2900 then it sounds like you forgot to subtract the resistor, though it is included in your equation here in the forum.
    Thanks for asking about checking Cyton Impedance outside of the GUI. Check your program to see if you are indeed subtracting 2200, and let us know the result.
    edited January 2020
    Richard, so regardless of his subtraction value for the series resistor, 640K is way different than what he is getting. My hunch is that there is something odd going on with the "...1.0e-6 / 6.0e-9..." part of the expression. Looking at the code segment on Github,
    (Lines 150+), the array variable, "dataProcessing.data_std_uV[Ichan]" is multiplied by 1 microvolt (1.0e-6). Yet why is the variable named uV? I guess it must be actually Volts and not microvolts? The result of (sqrt(2.0)*dataProcessing.data_std_uV[Ichan]*1.0e-6) is then divided by leadoff drive 'amps' (cyton.get_leadOffDrive_amps()). Six nanoamps, (6.0e-9).
    It's possible that there are some layers of DSP filters that precede this stage of calculation in the GUI. Such as to remove mains noise, and even zero in (via bandpass filter) on the particular frequency output by the impedance measurement generator in the chip. Which I think I recall is 31 Hz.
    Looking again at Hiroki's expression, I don't think it has the correct parentheses grouping to control order of expression evaluation. The GUI sources does this:
    float impedance = (sqrt(2.0)*dataProcessing.data_std_uV[Ichan]*1.0e-6) / cyton.get_leadOffDrive_amps();
  • Hello,
    The source code of GUI:
    My python source code is:
    impedances = [ sqrt( 2.0 ) * (-v) * 1.0e-6 / 6.0e-9 - 2200 for v in sample.channel_data ]
    I think the Python code is as same as the GUI's code.
    Must be some code outside of this, that is affecting the computation. I mentioned bandpass filtering.
    Remember, the only important amplitude being measured, is at 31.2 Hz. Any other EEG frequency components would contaminate that.
    Are you sure that data_std_uV[Ichan] is the same as sample.channel_data. This calculation should be looking at amplitude or power of the EEG, on that channel. Not raw sample values, which go above and below the Y axis. sample.channel_data looks like a raw sample to me.
    "Not raw sample values, which go above and below the Y axis." This is of course assuming you've done a high pass filter at say .5 Hz, to cut off the DC offset.
  • Oh, yes, sample.channel_data is raw data, not band-pass filtered.
    I use cyton.py to read EEG data, so sample.channel_data is a list of raw amplitudes at unit uV.
    I'll add a narrow band-pass filter (31.2Hz +- some Hz) to my code and try again.
    The expression in the GUI is NOT using raw EEG data samples. There is some preceding step in the GUI source code that you are missing that computes data_std_uV from the raw data. That is my guess. You CANNOT feed this impedance expression raw EEG values. Because raw EEG samples are both above and below the Y axis (plus and minus microvolts), the output will just be nonsense. The input to this expression is some type of estimate of power or amplitude of the overall EEG. Possibly related to what is displayed on the time series graph, showing the small numeric text on each channel, with the approximate rms power.
    Additionally, any sort of amplitude or power estimate would include some type of time averaging, running average. So that instantaneous small peaks are smoothed out.
  • Now I'm trying this problem. I think we can get the impedance data after executing config_board('z 1 0 1 Z') or config_board('z 4 1 0 Z') on Python. And I used this calculation "data2=data*4.5/ float((math.pow(2, 23) - 1)) / (24) * 1000000", which commented on brainflow's slack.
    But the impedance was different from OpenBCI_GUI. Could someone help me?
    @Hiroki
    Did you solve this problem??
  • Hi,
    When using the impedance settings to stream impedance values from the channels, should the raw output be multiplied by anything (similarly to the EEG and ADS scale factors) to get the Ohm values?
    Zach
    @ZachNairac, hi.
    Merged your new thread into this existing one on the same topic. See previous comments.
  • XaviWhiteXaviWhite Tarragona
    Good morning,
    I've been programming a little file to export the data impedance gived in channel 2 in which I'm trying to measure a 2.2kohm resistor. The problem is that checking with the OpenbciGUI shows me the correct resistance I'm measuring but I don't know if I'm doing correctly with the python file because I can not export the correct value. Any help, please?
    ThankU:)
    @XaviWhite, hi.
    I merged your comment into this existing thread, please see previous comments.
    See this section of the GUI code,
  • I'm still struggling with the impedance issue.
    The impedance values are still not accurate and always come out around 2.4. In addition, the value changes very little when the EEG machine is put on and off. Has anyone had any success with impedance measurements? Also, does anyone have any advice?
    Our code firstly changed setting by config.board('z X 1 0 Z'), and get the board data by get_board_data(). After that, we applied the 31Hz bandpass filter and calculated the standard deviation(data_std) for 1 second data. Then we calculated the impedance by using this formula.
    impedance = (sqrt(2.0)data_std_uV[X]1.0e-6) / BoardCytonConstants.leadOffDrive_amps(6.0e-9);
    impedance -= BoardCytonConstants.series_resistor_ohms(2200);
  • Sorry, my explanation is little wrong. I multiplied 1/1000 by the impedance. So the factor of BoardCytonConstants.series_resistor_ohms largely affected the value. So the value of the impedance is near 2.2~2.4.